A lot of the GC-MS analyses that are being posted online are totally inadequate and unprofessional and in some cases verging on useless.This is bad science. if you are going to go to the effort of running GC-MS then at least do it properly.
A crappy scan of a printout of a non subtracted spectrum with no TIC is not good enough.
posting crap data in the hope that some of the skilled people here will spend their time examining it is disrespectful and unhelpful. the old saying rubbish in rubbish out applies here.
GC-MS data should have the following data as a minimum.
Method:
1, The sample preparation technique whether any derivitization was used
2, The type of injection eg split (with split ratio) splitless on column PTV etc
3, The injector oven temperatures and carrier type and gas flow
4, The Mass spectrometer parameters, in particular the low mass cut off and scan range
Results:
1, The Total Ion Count TIC for the entire run
2, Any selected spectrum should be background subtracted and the background level selected just prior to the peak of interest, in Chemstation this is called the BSB
3, The TIC should be integrated and the integrated results provided to indicate area under each peak.
highly desirable is an extracted ion chromatogram for the 4 largest peaks for the peak of interest or the peak purity should be given, the shape of these extracted ion chromatograms gives a good indication whether these ions are associated with the peak, they also reveal post injector decomposition.
Instead of doing all the data processing on the capture system it is better to supply the raw data. If you are paying for analysis you should expect all the data to be supplied.
for example with chemstation this is the *.d folder with masslynx the *.raw and with finnigan the *.RAW should be supplied etc
the serial number of the MS can be easily removed from the data (in reality as the serial number can be spoofed nobody actually cares)
by supplying the full data then proper processing of the data becomes possible by third parties.
There is plenty of free software to deal with the data.
AMDIS http://chemdata.nist.gov/mass-spc/amdis/downloads/ is a very powerful free tool, you can thank the US chemical weapons verification team for it, for data analysis and can import the following file types,
* Agilent ChemStation (*.D)
* Agilent MS Engine (*.MS)
* Bruker (.MSF)
* Finnigan GCQ (*.MS)
* Finnigan INCON (*.MI)
* Finnigan ITDS (*.DAT)
* HP ChemStation (*.D)
* HP MS Engine (*.MS)
* INFICON GCMS (*.acq)
* JEOL/Shrader (*.lrp)
* Kratos Mach3 (*.run)
* MassLynx NT (*.*)
* Micromass (*.)
* mzXML/mzData (*.*) (new in NIST11)
* NetCDF (*.CDF)
* PE Turbo Mass (*.raw)
* Shimadzu MS (*.R##)
* Shrader/GCMate (*.lrp)
* Varian MS (*.MS)
* Varian SMS (*.sms)
* Varian XMS (*.xms)
* Xcalibur Raw (*.raw)
Wsearch32 is a good freeware program http://www.wsearch.com.au/wsearch32/wsearch32.htm
* Kratos DS90 , AEI DS55
* Kratos Mach3 (*.run)
* Agilent Chemstation (data.ms)
* Agilent LC/MS (msd1.ms)
* HP Pascal. (*.d)
* Varian Saturn (*.ms, *.sms)
* Finnigan ITS40,magnum (*.ms)
* Finnigan ICIS, ITDS (*.dat)
* Finnigan Xcalibur (*.raw)
* Shrader (*.lrp)
* JCAMP_DX (*.dx, *.jdx)
* netCDF (*.cdf)
* Joel (*.lrp)
* mzXML (*.xml, *.mzxml)
* Masslynx (*.raw)
* PE TurboMass (*.raw)
* Bruker XMASS
* Shimadzu (*.R##)
Thanks for listening.
Basic Good Laboratory Practice
it is good laboratory practice to run blanks (solvent only with no sample) before and after each sample, it is also good practice where references are available to run the reference immediately after the sample using the same conditions. The retention time should match within a 0.5 to 1 second. it is also good practice to run a standard mixture regularly. septa should be changed after no more than 200 injections, injector liners should be regularly changed, as should split vent filters, where dirty samples are used or excessive amounts of solvent is injected then the split vent lines and filters need regular cleaning
samples should not overload the column but should give peak heights at least 5 times noise, for example on a clean well set up HP 6890/5973 with a low bleed the background count is usually around 5-8000 counts peak heights should be 20 000 counts at a minimum. Sample injections should never exceed the volume of the injector liner when vaporised except in special cases. low MW solvents like methanol are the biggest offenders here.
samples should dissolve fully in the solvent, in the case of the sample not dissolving then this needs to be noted and attention drawn to this fact. where samples are filtered care needs to be taken to ensure analyte is not absorbed on the filter.
An Internal standard added as a spike to the sample are useful to ensure the sample is getting into the machine, where possible the internal standard should be of a known concentration and should have similar chemical properties to the sample to account for absorption on the injector and column, for example dichlorobenzylamine is a useful IS for phenethylamine type drugs. Naphthalene is quite good too but is inactive so doesn't reveal an active inlet.
The Mass spectrometer should be tuned at least weekly and quick tuned daily. The reduction in heavier mass abundance (502 with PFTBA) and peak shape indicates source clean is required. An air and water check should be regularly performed any increase in N2 and O2 should be investigated.
A crappy scan of a printout of a non subtracted spectrum with no TIC is not good enough.
posting crap data in the hope that some of the skilled people here will spend their time examining it is disrespectful and unhelpful. the old saying rubbish in rubbish out applies here.
GC-MS data should have the following data as a minimum.
Method:
1, The sample preparation technique whether any derivitization was used
2, The type of injection eg split (with split ratio) splitless on column PTV etc
3, The injector oven temperatures and carrier type and gas flow
4, The Mass spectrometer parameters, in particular the low mass cut off and scan range
Results:
1, The Total Ion Count TIC for the entire run
2, Any selected spectrum should be background subtracted and the background level selected just prior to the peak of interest, in Chemstation this is called the BSB
3, The TIC should be integrated and the integrated results provided to indicate area under each peak.
highly desirable is an extracted ion chromatogram for the 4 largest peaks for the peak of interest or the peak purity should be given, the shape of these extracted ion chromatograms gives a good indication whether these ions are associated with the peak, they also reveal post injector decomposition.
Instead of doing all the data processing on the capture system it is better to supply the raw data. If you are paying for analysis you should expect all the data to be supplied.
for example with chemstation this is the *.d folder with masslynx the *.raw and with finnigan the *.RAW should be supplied etc
the serial number of the MS can be easily removed from the data (in reality as the serial number can be spoofed nobody actually cares)
by supplying the full data then proper processing of the data becomes possible by third parties.
There is plenty of free software to deal with the data.
AMDIS http://chemdata.nist.gov/mass-spc/amdis/downloads/ is a very powerful free tool, you can thank the US chemical weapons verification team for it, for data analysis and can import the following file types,
* Agilent ChemStation (*.D)
* Agilent MS Engine (*.MS)
* Bruker (.MSF)
* Finnigan GCQ (*.MS)
* Finnigan INCON (*.MI)
* Finnigan ITDS (*.DAT)
* HP ChemStation (*.D)
* HP MS Engine (*.MS)
* INFICON GCMS (*.acq)
* JEOL/Shrader (*.lrp)
* Kratos Mach3 (*.run)
* MassLynx NT (*.*)
* Micromass (*.)
* mzXML/mzData (*.*) (new in NIST11)
* NetCDF (*.CDF)
* PE Turbo Mass (*.raw)
* Shimadzu MS (*.R##)
* Shrader/GCMate (*.lrp)
* Varian MS (*.MS)
* Varian SMS (*.sms)
* Varian XMS (*.xms)
* Xcalibur Raw (*.raw)
Wsearch32 is a good freeware program http://www.wsearch.com.au/wsearch32/wsearch32.htm
* Kratos DS90 , AEI DS55
* Kratos Mach3 (*.run)
* Agilent Chemstation (data.ms)
* Agilent LC/MS (msd1.ms)
* HP Pascal. (*.d)
* Varian Saturn (*.ms, *.sms)
* Finnigan ITS40,magnum (*.ms)
* Finnigan ICIS, ITDS (*.dat)
* Finnigan Xcalibur (*.raw)
* Shrader (*.lrp)
* JCAMP_DX (*.dx, *.jdx)
* netCDF (*.cdf)
* Joel (*.lrp)
* mzXML (*.xml, *.mzxml)
* Masslynx (*.raw)
* PE TurboMass (*.raw)
* Bruker XMASS
* Shimadzu (*.R##)
Thanks for listening.
Basic Good Laboratory Practice
it is good laboratory practice to run blanks (solvent only with no sample) before and after each sample, it is also good practice where references are available to run the reference immediately after the sample using the same conditions. The retention time should match within a 0.5 to 1 second. it is also good practice to run a standard mixture regularly. septa should be changed after no more than 200 injections, injector liners should be regularly changed, as should split vent filters, where dirty samples are used or excessive amounts of solvent is injected then the split vent lines and filters need regular cleaning
samples should not overload the column but should give peak heights at least 5 times noise, for example on a clean well set up HP 6890/5973 with a low bleed the background count is usually around 5-8000 counts peak heights should be 20 000 counts at a minimum. Sample injections should never exceed the volume of the injector liner when vaporised except in special cases. low MW solvents like methanol are the biggest offenders here.
samples should dissolve fully in the solvent, in the case of the sample not dissolving then this needs to be noted and attention drawn to this fact. where samples are filtered care needs to be taken to ensure analyte is not absorbed on the filter.
An Internal standard added as a spike to the sample are useful to ensure the sample is getting into the machine, where possible the internal standard should be of a known concentration and should have similar chemical properties to the sample to account for absorption on the injector and column, for example dichlorobenzylamine is a useful IS for phenethylamine type drugs. Naphthalene is quite good too but is inactive so doesn't reveal an active inlet.
The Mass spectrometer should be tuned at least weekly and quick tuned daily. The reduction in heavier mass abundance (502 with PFTBA) and peak shape indicates source clean is required. An air and water check should be regularly performed any increase in N2 and O2 should be investigated.